flo

Forum Replies Created

Viewing 5 posts - 1 through 5 (of 5 total)
  • Author
    Posts
  • in reply to: Frequency Resolution Spike Software #237317
    flo
    Participant

    Hi Andrew,

    thank for implementing more digits for frequency resolution in the latest software update

    BR
    Flo

    in reply to: Frequency Resolution Spike Software #237205
    flo
    Participant

    Hello, Andrew,

    Thank you very much for your answer. I look forward to the next software release.
    BR
    Flo

    in reply to: RBW and VBW #228762
    flo
    Participant

    Hi Justin,

    thanks for your answer.

    So turning off the image/spur rejection will not solve the problem?

    And if I set the VBW to 1/10 of the RBW and measure below 42MHz (I measure from 20 to 40kHz or 60 to 100kHz) I dont have to substract 3dB from the measurement.

    I will try it with the SA44B at first.

    in reply to: Noise measurement with Femto Preamplifier #198519
    flo
    Participant

    Hi likhil,

    connect the photodiode or DUT to a current/voltage amplifier (transimpedance amplifier) the output of the amplifier, should be ac or use a dc blocking element, connect to the Signalhound.

    etting Spike Software:
    Trace: Type: Average
    Acquisition: Video Units: Power
    Detector: Average
    Bandwith: RBW Shape: Flat Top
    RBW:10Hz
    VBW:10Hz
    REF Level: -50dBm
    Start: 1Hz
    Stop 200kHz
    Spur Reject: Off
    Gain/Atten: Auto

    BR
    Flo

    in reply to: Noise measurement with Femto Preamplifier #198510
    flo
    Participant

    Hi Justin,

    thanks for replying.
    Yes the gain is 1000 volts per amp.

    For example:
    -120dBm to V/rt Hz than using the gain 1000V/A to get A/rt Hz

    ???_(???/??)=?_???−10 log⁡(???) [-120-10log(10)]
    ?= (√(10^((?_???−30)/10)∗?))/???? (rt(10^((-130-30)/10)*50OHm)/1000V/A
    I = 5E-10A/rt Hz

    Is this the way to do?

    BR
    Flo