Reply To: Power Spectral Density Measurement Discrepancy?

Forums SA Series Discussions Power Spectral Density Measurement Discrepancy? Reply To: Power Spectral Density Measurement Discrepancy?

#59249
AndrewAndrew
Moderator

Hello Mike,

Can you try attenuating the input to ensure you are not in compression? Spike also sets the attenuator and pre-amp automatically based on reference level. With your ref offset active, set the reference level to +40 to see if this affects the measurement. I am concerned that the signal is possible being compressed in the Spike software.

Having said that…

I performed a similar measurement on both the Spike software and version 2.18. I was able to get similar results on both pieces of software. I have attached the measurements. In this case, I used a signal generator to create a wideband signal with an output power of -40dBm. Seeing a channel power output reading of -40dBm on both software versions lets me know I have configured the software correctly. If you have a generator, can you try a similar test?

Regards,
A.J.

Attachments:
You must be logged in to view attached files.